| 
 | 
	
 
 
95s红手报警 
A - PICK UP TUBE  IMPROPER POSITION         
 A管位置不对 
Adjust pump head to tube X.XX mm  Press BLOOD FLOW key to confirm 
改动了血泵管径,压血流量键确认 
AIR DETECTOR NOT ACTIVATED  Press AIR DETECTOR key         
空气监测没启动,压空气监测键 
AIR LEAKAGE  Check fluid tubes connections         
漏气,检查透析液管路接头 
Attach fluid tubes Press SELECT to abort the test         
接上透析液管路压箭头中止测试 
Attach fluid tubes Press SELECT to start disinf 
接上透析液管路压箭头开始消毒 
B/ACETATE - PICK UP TUBE IMPROPER POSITION         
B管位置不对 
BATTERY FAILURE Change battery or press SELECT         
电池失效,换电池或压箭头键 
BiCart emptied press SELECT to confirm         
干粉筒排空压键头确认 
BiCart HOLDER NOT CLOSED Close or check BiCart holder         
干粉支架未关好 
BiCart NOT ATTACHED TO HOLDER Attach or check BiCart         
干粉未接上支架,请接上。 
BLOOD IN PRIMING DETECTOR Function check stopped         
预冲探测器监测血或预冲探测器赃了。 
BLOOD LEAK DETECTOR FAILURE Clean and press BLOODLEAK key         
 漏血测试失败,清洗并压漏血键。 
BLOOD PUMP COVER NOT CLOSED Close blood pump cover         
血泵门未关 
BLOOD PUMP OVERLOAD IN FCH Check the pump and press SELECT         
自检时血泵卡死,检查血泵压箭头。 
BLOOD PUMP OVERLOAD Press BLOOD PUMP key         
血泵卡死,压血泵键。 
BLOOD PUMP STOP TIME EXPIRED Start Blood pump         
血泵停止时间过长,启动血泵。 
BPM FAILURE Turn off BPM         
血压计坏了,关闭血压计。 
CHEM DIS FILLUP PHASE COMPLETED Place pick up tube in machine 
消毒液吸满了,将吸液管放入机器内 
CHEMICAL FILLUP NOT SUFFICIENT Check supply of disinfectant         
消毒液不对,检测消毒液 
Connect concentrate when time to prepare machine for treatment 
请接透析液 
DECALCIFICATION REQUIRED XX hours since last decalcific.         
需要脱钙 
Dialysis flow too low         
透析液流量太小 
Dialysis fluid not ready for treatment. Fluid ready in xxx s         
透析液未准备好 
DIALYSIS FLUID READY PRESS BYPASS KEY 
透析液浓度好了,压旁路键 
DISCONNECT VENOUS PRESSURE Press VENOUS key to confirm         
取下静脉压保护器,压静脉压键确认 
DISINFECTION REQUIRED xx days since last disinfection         
需要消毒 
EXTRA PRIMING VOLUME ACHIEVED         
预冲量到了 
FLUID BYPASS TIME EXPIRED Press BYPASS key         
旁路时间太长,压旁路键 
FLUID PATH OBSTRUCTION Check fluid tubes and dialyzer 
管路中有堵塞,检查管路和透析器 
FLUID TUBE SENSOR TEST FAILED Press SELECT to confirm 
旁路管位置感应器测试失败,压箭头确认 
FLUID TUBE SENSOR TEST Attach fluid tubes         
旁路管位置感应器,接上旁路 
FLUID TUBE SENSOR TEST Remove fluid tubes         
旁路管位置感应器,取下旁路 
FLUID TUBES IN SAFETY BYPASS Connect fluid tubes to dialyzer         
旁路管在机器上,接上透析器 
FLUID TUBES NOT IN SAFETY BYPASS         
旁路管不在机器上 
FUNCTION CHECK RESTARTED Please wait         
机器自检重新开始 
HEATER EFFICIENCY NOT SUFFICIENT FOR PROPER DISINFECTION         
热消毒功率不够,检查电压和保险 
HEPARIN PUMP OVERLOAD Check heparin syringe and line         
肝素泵过载,检查针筒到底和管路夹死 
HEPARIN SET TO ZERO Press HEPARIN key to confirm         
肝素设定为零,压肝素键确认 
High conductivity not obtained  Check concentrate         
高电导达不到,检查A液 
HIGH INLET WATER COND. x.xmS/cm Check inlet water         
进水电导太高 
HIGH MONITOR TEMPERATURE ALARM Turn off the monitor         
机器温度高温报警,关上机器 
IMPOSSIBLE TIME OR UF SETTING Adjust TIME or UF VOLUME         
时间或脱水量设定不对,调节时间或脱水量 
INCORRECT ACIDIC CONCENTRATE Check concentrate         
A液不对 
INCORRECT BICARBONATE CONC Check concentrate         
液不对 
INCORRECT CONCENTRATE Check concentrate         
A,B液都不对 
INCORRECT DIALYSIS FLUID COMPOSITION         
透析液的成分不对 
INSUFFICIENT INLET PRESSURE Please check water supply         
进水压不够 
LOW BATTERY POWER Machine turns off within one minute         
后备电池电量太低,1分钟后关机 
NEGATIVE UF RATE Check and press UF volume         
负超滤率,检查脱水量设定和压该键 
NO BACKFILTRATION WARNING Press TMP to confirm         
没有反超报警,压TMP确认 
NO BPM VALUES Check BPM cuff         
压计没压力,检查绣袋 
NO CUFF ATTACHED Attach or check BPM cuff         
血压计绣袋没接 
ONLY BLOOD MONITOR IS RUNNING Press SELECT to start fluid part         
只有血路部分在运行,压箭头启动水路 
DLEASE WAIT 30 SEC. TO BYPASS DIALYSIS FLUID         
等30秒,旁路才好 
POWER FAILURE, BATTERY OPERATION Battery operated for XX minutes         
无交流电,后备电池维持了xx分钟 
READY FOR VENOUS PRESSURE TEST Press VENOUS key to start         
准备测静脉压,压静脉压键 
RECOVERED FROM POWER FAILURE Press TIME key to confirm         
非正常开关机,压TIME键确认 
Remove fluid tubes Press SELECT to abort the test         
旁路管被取下,压箭头取消测试 
SAGS TEST FAILED         
AGS开关测试错误 
SET DIALYSIS FLOW NOT REACHED Readjust flow or call service         
设定的透析液流量达不到,重新调节 
TMP LIMITS TOO WIDE Press TMP key to adjust         
TMP跨膜压范围太宽,压TMP调节 
To empty BiCart, press SELECT         
排空干粉筒,压箭头 
To end treatment, press isol UFand zero time         
治疗结束,压单超键 
To obtain total UF volume ISOL UF must be started         
使用单超,才能达到总的脱水量 
To start the air detector test:Fill the venous drip chamber         
装上静脉壶,开始预冲 
TOO HIGH TMP REQUIRED Increase time or decrease UF volume         
TMP跨膜压太高,增加时间或减少脱水量 
TREATMENT STARTED NO RESET MADE Press SELECT if OK, else RESET         
治疗模式不是接台模式,可以压箭头,否则重新接台 
UF RATE LIMIT REACHED Re-adjust TIME or UF VOLUME         
超滤率超极限,调节时间或脱水量 
UF RATE LOWER THAN MINIMUM SET Adjust TIME, UF VOLUME or MIN UF         
超滤率小于最小超滤率,检查时间,脱水量,最小超滤率 
UF STOP TIME EXPIRED Press START/STOP key         
脱水停止时间到了,压START/STOP键 
UF VOLUME SET TO ZERO Press UF VOLUME key to confirm         
零脱水设定,压UF VOLUME 键 
VENOUS LOW LIMIT BELOW +10 mmHg Press VENOUS key to adjust         
静脉下限低于+10压静脉压调节 
VENOUS PRESSURE LIMITS TOO WIDE Press VENOUS key to adjust         
静脉范围太宽,压静脉压调节 
VENOUS PRESSURE TEST INTERRUPTED Press SELECT to confirm         
静脉压测试中断压箭头确认 
Wrong disinfectant agent Press SELECT to restart disinf.         
错误消毒剂,压箭头重新开始消毒 
 
 |   
 
 
 
 |