F14 pos. Pressure
Zero point change after closing of solenoid valve V43 (20 sec).
Standard: P-Dial. not equal to –125 to 55 mmHg.
HDF option: P-Dial. not equal to –125 to 60 mmHg.
– The solenoid valve V24 is not closed.
The balancing chamber is switched to passage during this test sequence. V24, V26 and
V43 are closed; V24B is open.
F14 Shortly before the end of the mandatory rinse in Dis I to V.
The hydraulics test has not been completed correctly, possibly caused
by flow problems.